86108B Clk Rec. Setup

Instrument
DCA-X
N109X
Flex Apps:
FlexDCA

Use this Setup dialog to configure N1060A and 86108A/B precision waveform analyzer modules.

Clock Recovery

Use the Clock Recovery tab to a configure clock recovery for an N1060A and 86108A/B precision waveform analyzer. For all other clock recovery modules use this dialog box. These two dialog boxes are identical except to the channel and precision timebase tabs that are included for the N1060A and 86108A/B.

When the precision timebase is configured to use an external reference in the dialog box's Precision Timebase tab, the clock data recovery circuitry is disabled.

An autoscale does not cause clock recovery to re-establish lock.

Input Fields

Use the Input fields to select the input Source for clock recovery. Enter the Symbol Rate of the signal directly, or click Select From List to select from a list. Since the loop bandwidth capture range is wider than the selected symbol rate, it is possible for the module to lock on a signal that is higher than the selected value. For example, if you select a 2.48832 GBd symbol rate but the signal is actually 2.5 GBd, the module may still lock on the signal.

Input Spread Spectrum Clock Field

FlexDCA uses the average symbol rate to recovery the clock. However, when performing clock recovery on a Spread Spectrum Clock (SSC) signal, using the nominal symbol rate may be required. When Spread Spectrum Clock is selected, the nominal rate is used.

Clock Recovery Fields

Click the Lock button at the top of the Clock Recovery field to lock the clock recovery. The Loop Bandwidth fields are used to configure the loop bandwidth.

Select Auto Relock to enable automatic signal re-locking while circuit probing. This feature frees you to concentrate on handling the probe rather than locking the instrument.

Loop Bandwidth (LBW)

Use the Loop Bandwidth fields to configure the bandwidth of the clock recovery's phase locked loop. Because jitter is measured relative to the clock, jitter that is present on both the data and the recovered clock cannot be detected by the instrument. Jitter that is within the clock recovery module's loop bandwidth appears on the recovered clock. Jitter that is outside of the loop bandwidth does not appear on the recovered clock and can be detected on the data. For this reason, you can select the loop bandwidth for the clock recovery module. The available loop bandwidth settings depend on the module as shown in the following table. Fixed values are entered in Hertz; rate dependent values are expressed as a symbol rate ratio (LBW = rate/n). The selected bandwidth applies both to the recovered clock and to the front-panel Recovered Clk Output. You can also adjust the equalization level on the PLL loop characteristics for clock recovery as explained in Equalization Level later in this topic.

Available Loop Bandwidth Definitions by Module
Module Available Loop Bandwidth Setting
Fixed Rate Dependent Low
(270 kHz)
High
(1.5 MHz)
N1060A
N1078A
N1077A
N1076B
N1076A
N4877A
83496A/B Option 300
83496A/B Standard

Output Fields

Front Panel Output

Use the Output panel to set the output clock divide ratios. Select Auto Divide Ratio for automatic selection of the clock divide ratio that is applied to the front-panel recovered clock output. Available divide ratios vary from 1:1 to 1:32 depending on the symbol rate and loop bandwidth settings. With manual selection, select the divide ratio from available settings in the drop-down list.

Trigger on Data

Select Trigger on Data when triggering on data instead of recovering the clock. The data is passed to the rear-panel via an AC coupled limit amplifier. Symbol rate and loop bandwidth settings are not adjustable. Triggering on the data is similar to not having a clock recovery module and using a splitter to divide the input signal and routing one portion to the front-panel trigger input. While triggering on the data provides a quick method of checking the data waveform, be aware that the displayed response may be missing 75% of the data patterns. This happens because triggering occurs at either the 1-to-0 or 0-to-1 data transition, and, as a result, the instrument will only be triggered when a transition occurs in the data instead of triggering on any symbol period as with a clock. For example, consider a pattern such as 1111000. Only the transition between the 1-to-0 triggers the instrument. Transitions between adjacent lows and highs will not cause trigger events. The recovered clock will still be available at the front panel Recovered Clock Out even when set to trigger on the data.

JSA Jitter Spectrum Analysis

Click Setup to configure jitter spectrum analysis, which is available in Oscilloscope, Eye/Mask, and Jitter Modes.

Requires an N1060A-JSA, N107x-series clock recovery module, or 86108A/B-JSA module (Option JSA. Jitter Spectrum Analysis) to be installed in the oscilloscope.

Advanced Settings

Click the expander arrow at the bottom of the dialog to view advanced settings, which provide additional control of the clock recovery loop dynamics. The purpose of clock recovery is to extract a clock from the incoming data signal and to provide the DCA with a trigger that is synchronous with the data signal. The clock recovery loop bandwidth primarily determines how well the recovered clock tracks low-frequency jitter on the input signal. Some signals have very large low-frequency jitter from either extremely dirty clocks or intentional modulated clocks such as found in SSC (spread spectrum clocking). In this case the 86108A clock recovery system provides additional control of the loop dynamics allowing you to select the loop's Type-2 Transition Frequency.

The Type-2 Transition Frequency indicates the frequency below which the second integrator in the loop starts to provide extra gain. Increasing this frequency provides additional loop gain and improves the tracking of the loop. The following figure shows the jitter multiplier as a function of jitter frequency for a loop-bandwidth setting of 5 MHz and various settings of transition frequency. This multiplier is the magnitude of the observed jitter transfer function (OJTF). This additional tracking also increases the peaking in the closed-loop jitter transfer function (JTF). Calculated values of the peaking are provided in the user interface next to the available selections for Type-2 Transition Frequency.

Figure. OJTF for 5 MHz LBW vs. Type-2 Transition Frequency

The Type-2 Transition Frequency is automatically coupled to CDR loop bandwidth in normal operation (Auto is selected) and provides the desired loop characteristic for most measurements. The current setting is displayed in the Type 2 Loop Transition Frequency (Peaking) list. To override the automatic selection, unselect Auto and select an available Type-2 Transition Frequency value:

  • 12 kHz (available for all loop bandwidths)
  • 280 kHz (available for loop bandwidths > 600 kHz)
  • 640 kHz (available for loop bandwidths > 1.6 MHz)
  • 1.3 MHz (available for loop bandwidths > 4.5 MHz)

Equalization Level

The Equalization Level setting fine tunes the equalization level on the PLL loop characteristics for clock recovery. The default value is 0%. The range can be set from 0% to 100% in increments of 1%.

Loop Bandwidth Tuning

When locking clock recovery, Loop Bandwidth Tuning increases the accuracy of the Target LBW (Loop Bandwidwith) setting. Loop bandwidth tuning calculates the loop gain on live input data to arrive at an improved loop bandwidth setting. This feature is recommended when performing measurements that are sensitive to loop bandwidth, such as TDECQ.

When Tune on Relock is selected, loop bandwidth tuning is selected to be appended to the normal clock recovery relock algorithm. In other words, when you select Tune on Relock, loop bandwidth tuning is not applied until the next time that clock recovery is relocked. Click the Clock Recovery Lock button at the top of the dialog to initiate a relock. Enabling this setting increases the time needed for the clock recovery to lock. For example, it could add an additional 3 or 4 seconds when locking the clock recovery. By default, Tune on Relock is turned off.

In the following situations, Loop Bandwidth Tuning may not result in reliable adjustment in which case, turn off this feature:

  • Degraded input signals
  • Low-amplitude input signals

When clock recovery is locked, changes to any of the following loop bandwidth settings will invalidate the loop bandwidth tuning and the tuning status will become untuned:

  • Target LBW setting
  • Loop characteristics (peaking)
  • Equalization Level setting
  • Enable High Gain setting (N1078A DCA-M with optical input data)

Presets

Presets allow you to save your dialog settings to a setup file. Recall a preset to instantly configure your clock recovery to your specification. You can save as many presets as you need. The Preset list shows all of the factory provided presets as well as any that you have created. If you scroll to the end of the list and click the <Edit List> entry, the Edit Clock Recovery Presets List dialog opens which allows you to reorder, delete, or rename items in the list. In the list, the names of factory provided presets includes the text (factory) which indicates that the entry cannot be edited. The following factory presets are included in the list:

  • CEI-25G-LR
  • CEI-28G-SR
  • CEI-28G-VSR/MR
  • IEEE 10G-KR4_CR4
  • IEEE 10G-KR_40G-KR4
  • IEEE 40G-CR4_100G-CR10
  • Infiniiband EDR
  • PCI Express Gen 3
  • SATA Gen 1
  • SATA Gen 2
  • SATA Gen 3
  • USB 3.0 Gen 1
  • USB 3.0 Gen 2
  • To save your settings, click the button.